  • cruizer
    replied
    Sorta, kinda, really the machine requires to be loaded via a load bank like a Miller 750/Lincoln 750 load bank to test the machine to data plate specifications.

    Then its meter/s tested to a calibrated volt/amp meter. Only then can it be certified, providing the meters are within 5% of the calibrated meters used.

    A screwed up machine will NEVER put out to data plate specs, but may be within meter to meter readings.

    I've calibrated literally thousands of machines

    Leave a comment:


  • walker
    replied
    We just used a calibrated amp clamp to measure output amps and compared it to the machine settings.

    Leave a comment:


  • M J Mauer
    replied
    Welcome to the forum.

    Leave a comment:


  • charmon
    started a topic Syncrowave 351

    Syncrowave 351

    OK heres the deal i have always used engine driven welders,now i work for a company and their safety and equipment audit is coming up and i was told that our welder(SYNCROWAVE 351) needed to be calibrated i have never heard of this can someone give me an idea where to start?